Re: New guy with "msd" flow bench...problems
Posted: Thu Jan 12, 2012 6:00 pm
by coulterracn
If you decide to replace the orifice disk with some of Bruce's internal orifice plates, you will need them rated at 6" to work with the Dwyer incline manometer. Then use orifice plates rated at 28" on top of the bench for calibration. That's the setup I used when I had the Dwyer incline manometer on my bench.
Ray
Re: New guy with "msd" flow bench...problems
Posted: Fri Jan 13, 2012 11:47 am
by rangergt
Thanks, that makes sense. I've emailed Bruce.
Re: New guy with "msd" flow bench...problems
Posted: Fri Jan 13, 2012 2:46 pm
by rangergt
Ok after rereading more post I think I have an idea of whats going on and a plan. Using this
spreadsheet I could make up may own orfice plates for 6" DP for inside the flowbench that should be within reason and then use Bruce's PAP on the top to calibrate from there. Is my line of thinking correct?
Re: New guy with "msd" flow bench...problems
Posted: Fri Jan 13, 2012 5:22 pm
by Brucepts
Yes, that will work.
Re: New guy with "msd" flow bench...problems
Posted: Sun Jan 15, 2012 11:23 am
by pablo
I would get rid of the stainless mixing bowl's used for the intake and exhaust
control valves.
Fabricate something like this. This is from the Bill Jones website
Re: New guy with "msd" flow bench...problems
Posted: Sun Jan 15, 2012 2:40 pm
by rangergt
Yes, that is another thing that always bugged me about this flowbench. I was also planning to upgrade them as well. Thanks for the input.
Re: New guy with "msd" flow bench...problems
Posted: Sun Jan 15, 2012 3:42 pm
by 1960FL
Just for information about sharp edge orifice plates Bruce can make anything you want but sometimes it is unclear to how they work base on location. An example will be a 1.50” ID sharp edge plate this plate will flow the following and depending on where you put it will dictate its flow and use.
6” DP 74.7 CFM As an internal plat using a STD Dwyer incline
16” DP 122 CFM As an internal plate using the PTSDM
28” DP 143.1CFM As a calibration plate on top of the bench
It is all the same plate just what it flows at different pressure differentials. This was all assuming std air and .62 CD
Rick
Re: New guy with "msd" flow bench...problems
Posted: Sun Jan 15, 2012 9:12 pm
by larrycavan
That bench is fixable. Just how you go about it depends on how much flow you need.
http://www.cavanaughracing.com grab the phone number and call me during the week. About a 10 minute conversation should get you in good shape.
Re: New guy with "msd" flow bench...problems
Posted: Sun Jan 15, 2012 11:07 pm
by 1960FL
Thank you Larry you are the one that knows this bench.
Rick
Re: New guy with "msd" flow bench...problems
Posted: Tue Jan 24, 2012 12:40 am
by rangergt
larrycavan wrote:That bench is fixable. Just how you go about it depends on how much flow you need.
http://www.cavanaughracing.com grab the phone number and call me during the week. About a 10 minute conversation should get you in good shape.
Thanks a lot. I was away on business for a week with no computer so I'll give you a call this week.
Thanks for the info as well Rick.
Paul