Old Grey wrote:Just ignore what I'm building, I was just wondering how the PTS does such small readings with such a minuscule pressure ∆.
There are several reasons for that.
The first, as Bruce has already mentioned is that the whole of the entire range of a 16 inch pressure transducer is utilised, rather than a fraction of the range of a 40 inch range pressure transducer.
These pressure transducers all operate off a five volt dc supply, and the output voltage swing is specified from typically 0.5 volts up to 4.5 volts at full rated pressure. The output voltage from the transducer is unable to swing over the entire five volt supply range.
Typical analog to digital converers have an input voltage measurement range that extends right from zero up to typically five, or sometimes ten volts full scale.
If you hook up one of these pressure transducer directly to an analog to digital converter, (as FP have done) you will only be able to measure at most the four volt output swing that comes directly out of the transducer.
The PTS digital manometer has an amplifier located between the pressure transducer and the analog to digital converter. This amplifier causes a zero to 16 inch pressure measurement range to drive the analog to digital converter over the entire ten bit measurement span, from zero to ten volts.
With the FP, 16 inches input pressure is only 40% of the 40 inch transducer measurement span.
Not only that, the 0.5 to 4.5 volt output swing from the transducer is only 80% of the analog to digital converter's input measurement span.
The result is that the analog to digital converter can only actually read 80% of the 40% of the specified transducer pressure range.
A ten bit converter will have 1024 steps of resolution.
The FP will only be able to access 32% of those 1024 steps, or 328 steps of resolutionfor for 0 to 16 inches of input pressure measurement.
This makes the resolution of the PTS digital manometer which uses the entire 1024 measurement steps for 0 to 16 inches, THREE TIMES HIGHER, even before you start applying any averaging software to the readings.
The real power of very fine pressure resolution, especially at the low pressure end is in the averaging software.
But if you cripple the pressure measurement system resolution right at the measurement point, as FP appear to have done, you can never regain that loss with software.