by 84-1074663779 » Sun Sep 18, 2005 7:28 pm
What happens is that as the fluid rises in the sloping manometer tube, the fluid level falls in the reservoir.
Suppose your design manometer rise is fifteen inches, and the fluid level drops one quarter inch in the reservoir. The REAL manometer rise will therefore be fifteen and a quarter inches. Only a small error, but still an error.
The exact rate of fluid rise and fall is proportional to the surface areas of tube and reservoir. You don't need a lot of fluid volume in the reservoir, just as large a surface area as possible, that is the trick.
For a cylindrical reservoir, make it reasonably long and mount it horizontal, and half fill it so it has the maximum possible exposed surface area. Setting the fluid level to zero can be frustrating unless an easy method of adjustment has been provided for. I have used a reasonably large screw in plug (brass 1/2 BSP pipe plug) as a fill port in the top of the reservoir. I then use a small syringe to add/remove fluid to get the zero exact.
A further advantage of having a very large surface area reservoir is that it is much less sensitive to slight fluid loss, the zero point is very stable and easy to set.
Another simple way might be to raise or lower the whole reservoir mechanically to adjust the zero point. Some sort of hinge and set screw maybe ?
Getting a ratio of 1000:1 fluid relative surface areas between reservoir and manometer tube is fairly easy. An 0.1% error can probably be ignored.
Use a similar volume dry reservoir at the top so an over pressure accident will allow all the fluid to just drain back by itself. Both reservoirs can be mounted behind the manometer board out of sight for a neat appearance.