by Tony » Wed Jul 22, 2009 6:02 am
Once you get your head around the basic concepts, it will all suddenly fall into place for you.
You will have a kind of "Eureka moment" where the light of understanding suddenly turns on.
The basic idea is that your sloping manometer can be calibrated in percentage of flow, and that very same percentage flow scale will work with any orifice, or orifice pressure drop that you finally decide to use.
The trick is, to decide on an appropriate orifice size and pressure drop at the 100% flow point. There is no single correct or "best" pressure, it is a compromise.
You can run a smaller orifice that has a higher pressure drop at the required 100% calibration flow. Suppose you decide to use an orifice that dropped twenty inches of water at xxx CFM. The sloping manometer is then set so that the high end is twenty inches above the low end. So at xxx CFM the water rise is twenty inches. That corresponds to xxx CFM being 100% on your flow scale.
Or you can run a larger orifice that drops only five inches water at the same xxx CFM. In that case, the high end of your sloping manometer would only be five inches above the low end. The 100% flow marking on your scale still corresponds to the exact same xxx CFM, and everything still works the same.
But.......... There are some things to think about here.
First thing is that a very low manometer slope is much more prone to error if the manometer is not exactly straight or level. A completely vertical manometer will read right, even if the manometer tube is a bit off vertical, or the manometer tube a bit curved. A nearly horizontal manometer needs a LOT more care in construction and levelling.
Second thing is, a large orifice running at a very low pressure drop will be a LOT more intolerant of any up stream air turbulence, or slight imperfections in your flow bench.
A small orifice running at a high differential pressure drop will have a lot more violence and fury right at the orifice, and will read more accurately even if the entering up stream air is a bit turbulent, or your bench internals have flow instability or imperfections.
So basically it is a trade off. A higher design orifice pressure drop across your measurement orifice generally leads to a better performing bench.
But unfortunately this also makes your air blower work a lot harder, and it will ultimately reduce the maximum flat out available flow volume, due to the extra back pressure load placed on your air blower.
My advice would be to use the highest orifice pressure drop you, (and your air blower) can live with, depending on how much blower power you have, and what you intend to test with your bench.
Also known as the infamous "Warpspeed" on some other Forums.