by Terry_Zakis » Thu May 26, 2005 8:37 pm
Hello Larrycavan,
Apoligies up front. This is long.
I quickly found some infromation on turndown at:
Flowmeters and Turndown Ratio
An introduction to turndown ratio for flow measurement devices as orifices, venturi meter etc.
Turndown ratio is often used to compare the span - the range - of flow measurement devices.
Turndown Ratio
Turndown ratio can be expressed as:
TR = qmax / qmin (1)
where
TR = Turndown Ratio
qmax = maximum flow
qmin = minimum flow
Maximum and minimum flow is stated within a specified accuracy and repeatability for the device.
Example - Turndown Ratio for an Orifice Meter
The turndown ratio - TR - for an orifice meter with maximum flow of 12 kg/s and a minimum flow of 3 kg/s can be calculated as:
TR = (12 kg/s) / (3 kg/s) = 4 - normally expressed as turndown ratio of 4:1
This is a typical turndown ratio for a orifice plate. In general a orifice plates has turndown ratio between 3:1 and 5:1.
Turndown Ratio and Measured Signal
In a flow meter based on the orifice or venturi principle, the differential pressure upstream and downstream of an obstruction in the flow is used to indicate the flow. According the Bernoulli Equation the differential pressure increases with the square of flow velocity. A large turndown ratio will cramp the measurement signal at low flow rate.
Here is another source for some good reading on measuring flow with differential pressure meters, at Omega Instruments:
The orifice meters that I purchased for my "flow provers" have an integral orifice and orifice carrier, machined into one assembly, for being inserted between piping flanges. The orifices were set up based on the ISO Code 5167-1998, which parallels the ASME MFC-3M-1989, for flow within closed conduit (pipes).
Correction, please note that my orifices were calibrated at 70% of the maximum flow range.
My plan has been to use one wall of my shop, 26-feet long, to run a set of flow provers, which I could run from from my bench through, and use the provers as a check on the accuracy of the benches that I build/develop. For my provers I have four orifice meters, one 3-inch turbine meter, a 4-inch LFE, and two 2-inch LFE's.
For the orifice meter provers I have 4 of these setup's : one 6-inch, 0.725 Beta, a 4-inch at 0.725 Beta, a second 4-inch at 0.600 Beta, and finally a 3-inch at 0.600 Beta ratio. All were set up with a dual range calibration, that is, there's a calibration point for 0-20" of water colum, and a calibration point for 0-10" of water column differential across the orifices.
According to the ISO 5167 code, the calibration point is at 70% of the maximum flow. Maximum flow would be at the respective 20, or 10 inches of w.c. The 4:1 turndown ratio was used to determine the mimimum flow range for each of the orifice meters. Based on this information, each of the orifices should be able to measure flow within an uncertainty band of +/- 1%, or less. I made sure to observe all of the rules for lenghts of straight pipe runs, upstream and downstream of the orifice meters, and on the direction coming from the bench, the 6-inch line has a flow straightener that I fabricated out of 1" and 1/2" copper tubing, about 2-foot long, inserted into the line. This straightener will get rid of any rotational flow caused by the two "out of plane" elbows upstream.
Anyway, with great expense and time, I'm looking at just under +/-1% uncertainty, after you take into consideration the orifices themselves, the differential pressure transmitters, static pressure transmitters, (Honeywell Smart Transmitters), the thermocouples and flow computers. But this accuracy would only apply between my Max and Min flow ranges, within the 4:1 turndown ratio. Yes, I could measure flow below the Min level, but with the understanding that my uncertainty would increase, and therefore accuracy would decrease.
This is one of the reaons that I sprung for the new 4-inch LFE, because they are very accurate and have a much better turndown capability.
I didn't follow your statement that you perform your calibrations at 4 different flow levels. I wasn't sure that was possible, unless your orifice was calibrated at multiple flow levels to some kind of primary calbration standard?
Well, I hope that some of this has made sense. Hopefully without offending anyone on the excellent forum, I've contended for a while that I just don't belive the +/- 0.5% accuracy numbers that are so frequently stated capable of internal orifice benches. Yes, I agree that absolute accuracy is not paramount, but that repeatability is. Still, this is the long route which I've been taking.
I still think that Tony is quite brilliant on this stuff, as I've always been very impressed with reading his work. So I really do belive that the internal orifice bench, with large plenum, and laminar flow approach, could produce outstanding accuracy numbers. For my own learning, I'd just want to make some comparisons between a few differnt bench designs.
Best Regards,
Terry Terezakis
North Hatfield, MA USA