Your partly right. All else being equal, changing the angle changes the pressure at which your scale would read 100% and hence the calibration of your flowbench. So, if your change the angle of the scale you don't need to redo the scale but you do have to recalculate what 100% on your scale means.
The important thing to realize is that 100% on your scale in complete arbitrary, you choose the cfm value you want for 100% or, you choose the height and length of your scale and calibrate 100% accordingly.
For example: let's say you've got a 1.5" measurement orifice and you'd like 100% on your scale to be 100 cfm, that would require you to angle your scale so that it reads 100% at approximately 10.75" WC pressure difference exactly what angle is required to acheive that will depend on the construction details of the manometer and the density of the fluid you're using.
Now, let's say we switch to a 2.00" measurement orifice assuming the manometer is set up as above, a 100% reading for this orifice would be 178 cfm, the scale would have to be similarly calibrated for each of your orifices.