On Monday I was finally able to run some initial tests with my MSD based bench. Before I give the results, here's a quick profile of my bench.
* six 16-1234 motors
* 1 meter inclined manometer scale
* 12" rise
* orifice disc with 8 measurement ranges (from 29 cfm-
495cfm)
* cabinet leakage at 28" using the 29cfm orifice was .5 cfm
for the intake. haven't done leakage testing yet for exhaust
Using a 160cfm calibration plate that Bruce made for me, my bench at 28" measured 155cfm. That was using an orifice on my disc that should flow 189cfm at 100%. I ran several tests which were all within 1cfm.
I'm pretty happy with this result. However, during the testing I made a few discoveries.
1. Unless I have something really out of wack in the bench, six of these motors will not get me close to my goal of being able to measure 400cfm. As I recall Mousehouse1 had trouble getting to 400cfm with 10 of these motors. I 'm going to address this in a separate post.
2. I do not believe just using the control valves will give me enough precise control without adding some type of electronic means to throttle back at least one motor.
3. When I ran a test using the calibration plate on the next largest orifice on my disc, the results were no where close. I used an orifice on my disc that should flow 300cfm at 100% at 28". Using the same calibration plate the bench measured about 43cfm. By the way, I had no trouble getting the depression to 28" which sort of surprised me.
There has been much published and stated that you need to always select an orifice size that will put the inclined manometer in the upper portion of the scale, 60% or higher.
Has anyone else had similar results when the likely result would fall below 60% of the inclined scale?
So at this point, I'm not sure if I simply just need more horsepower so I can measure flow to 400cfm, or if I've got other issues also.
As always, I would appreciate any comments, suggestions, observations.
Thanks.
Dom
ps...this is fun but frustrating