by larrycavan » Mon Mar 20, 2006 10:18 pm
Increases in Verticle Drop will extend the range capacity. You're giving yourself more measurable flow by increasing the Verticle drop.
Whatever your verticle drop is, that's your max measurable pressure drop. After that, you run out of scale. Pressure Drop is what you're measuring. Pressure differential from top to bottom side of the orifice inside the bench.
You can measure it with a U tube but it's harder to read small changes. The inclined functions exactly like a U tube, it just has the scale arranged in % arrangement. It reads % of the verticel pressure drop. Nothing more, nothing less.
For a 2.17" diameter orifice with a Cd of .62 and a 9" verticle drop on your inclined manometer using 1.0SG fluid [or your scale being compensated for some other weight fluid] you would have a measurable flow of 191.55 CFM flow range.
For the same orifice but a 12" drop it would be 228.18 CFM.
Now, the tricky part is...Does the orifice have an actual .62 Cd?
You're proving that by comparing it with a calibration orifice of a known flow value at a known test pressure.
Without the known calibration to compare the orifice to, you are going to drive yourself crazy.
You first have to know the construction of your manometer in order to rely on what it reads. There are 4 basic adjustments to the inclined manometer that can influence the readings.
1. fluid sg
2. verticle rise
3. inclined leg to well ratio
4. a properly laid out scale
If all 4 of those pieces of criteria are not correct or compensation in one because of a change in another has not been taken into account, you will get bogus readings.
Larry