I do not have any Clippard valves for resale.
What I am saying is this; we use a 16" Dp sensor and use a 16" Dp orifice plate or 100% of the sensor's range. If someone uses a 40" sensor and only uses 6,8,10, 16 Dp etc they are only using a percent of the possible range.
A sensor has a working range much like an inclined manometer does and you should try and stay in that range. The sensor's are way more sensitive than a water gauge though so they tend to work in the lower part of the scale. We see this with the 16" sensor when using a large plate internally ie 400cfm and trying to measure down to ~25cfm. Depending on your averaging and computer speed your numbers might be slow to change on screen.
We are using the full range of the sensor, imagine if you were only working in a percent of the total range of the sensor? In order to take full effect of a 40" sensor you would need a 40" Dp for the full range.
Currently, with the PTS DM you need to use a 16" Dp across the plate. The math within Flowsoft only works in this range, this was done for the best possible linearity across the working range. This means you need to have 16" more vacuum below your plate than above the plate for 100% across the plate.
It really can be confusing . . . and we are still learning what we can do with this DM.
PS: I'm also looking at 100" sensors for static pressure for those who want to test above 38". My PTS Bench design does not allow for testing above 38" due to it's basic simplistic design in construction.