by Terry_Zakis » Sun Dec 18, 2005 7:40 am
I know this thread is about leakage, but some interesting points have come up about orifice vs. LFE's, so here are a few points which may be obvious to some but not others.
All of the reasons cited here in support of an orifice for measurement, are also valid in industry. If manufactured properly and not affected by non-uniform, pulsing, or spiraling flows, the orifice can give very good, and repeatable measurements.
In custody transfer with orifice meters, use of calibration is only done to verify that the orifice was manufactured/assembled correctly. After that, the calculation of the flow doens't rely on the calibration at all.
The laminar flow elements are also outstanding devices, but do have their limitations as cited above from contamination. As opposed to orifices, the LFE's depend 100% on their calibration, so if you get passages damaged or plugged, then the calibration is no longer valid, and your results will be in error. On the other hand, all you'd have to do is inspect, and clean the orifice plate and you'd be back in bussiness.
I have one 4-inch, and two 2-inch Merriam LFE's, and plan to use them for calibration/verification of my orifice based flow provers (which are external to the bench).
I'm also interested in using the LFE's to baseline the performance of the orifice flow provers, which have known limitations from turndown. As flow drops off through an orifice, and is no longer turbulent (defined as having a Reynolds number +10,000), the Cd would no longer be valid, and could give erroneous results. I have a few Dart flow computers that I'm setting up, which by design, are dual-range, having inputs for two Cd's (design flow range and then a low flow range). So in theory, the same orifice can accomodate lower flows, with a different Cd, and higher uncertainty in the measurement.
I also read a post a few days ago, that supported use of orifice plates in pipes, but I'm not sure if they were outside of the bench.
The one thing that amazes me about reading Tony's work, is his persistence in having a large plenum prior to and after the orifice, or as much as can be accomodated. When you take a step back and think about it, it's quite brilliant.
The ASME flow codes will show you that your measurement uncertainty goes down, as you increase your Beta ratio (pipe inside diam / orifice diam). So what Tony has been doing and advocating is use of an orifice that has a much higher Beta ratio than one could practically obtain within a pipe.