I've gone through most of the posts here and there is a lot of good information..
A question that has come up is about calibration using the calibration plate on a SF110
According to here, the 1.874 / 0.3125 plate flows 153.2 CFM at 10 inches.. In the Superflow manual published in 2005, the manual states that the % manometer should read 78% at 10 inches on the 185 scale and that all calibration should be based on this..
Lets assume that my plate reads 78% at 25 degree delta (it don't but I want to assume that it's perfect).. OK, 78% of 185 is 144.3 and I need to correct for the 25 deg delta. (x0.981) so the calculated flow of the bench is 141.6
This means that my bench is reading 0.924 (141.6/153.2) of what is should be reading.
If I up my calibration factor from 78% to 84.4% (78/0.924) then when I have a perfect bench I should get 84.4%
185 x 0.844 x 0.981 = 153.2
Now I'm right on...
Seem like the manual is messed up.. I shoud be calibrating off 84.4 not 78..
Any clue?
TIA Bruce